I have a rather sturdy isolation transformer in my home lab and I just realized that the primary side earth is not only connected to the output side earth but also to the output side neutral. I don't understand why, but it explains some odd measurement results I've had when I assumed everything was isolated from earth.
Inside of the whole device, toroid is potted inside and halfway up outside. Fuse is in the rightmost switch.
Closeup of the connections corresponding to the schematic.
simulate this circuit – Schematic created using CircuitLab
The simplified schematic (without the filter and protection) shows where the input earth is connected to the chassis but also to the respective output earth and to the output neutral.
I could not find documentation for it. The label is old and faded, judging from the markings on the capacitors and power switch it was made no earlier than 1989.
PROD NO : 60065
LOW IMP TRANS ISOL
SERIAL NUMBER :
0750-UK 9003048
MADE IN SWEDEN
Why is it like that?
When is this ever useful? Since I got this from an electronics lab surplus auction I don't have the manual, but the label in the back says "LOW IMP TRANS ISOL". Have they just used it as a large line filter? It seems to me that with the current configuration it is as deadly as a direct connection to the outlet.
I thought that the point of an isolation transformer was to remove the earth ground potential or tie on the neutral wire. Even the venerable Olin was taken by surprise by a device such as mine and blew a fuse.
How should I modify it?
My primary goal is to isolate devices I want to measure with my oscilloscopes (DUT).
First of all I should probably keep the transformer chassis connected to the input earth and remove the link from the chassis to the output connectors.
But then I have at least two or three options:
Remove every earth wire on the output connectors, including the link C to the output neutral. This is how I actually expected the transformer to be wired.
Connect the output earths to each other but remove the link C. This seems pointless to me. The only thing it does is to make sure two or more different devices will have their earths connected, something you may or may not want but probably don't care about in this context.
Leave the outputs as they are, just remove the connection between input and output. The output neutral will then still be connected to the output earth, but will be floating with respect to the input side earth (the "real" earth).
If I understand some other answers here, option 1 is really the way to go, but I wonder if option 3 would give me any kind of benefits?