I have one of these power supplies:
I also have an MCR-4V data logger:
If I connect wires directly from the power supply to the data logger (red to red, black to black), then the voltage output on the power supply closely matches the measured voltage, e.g. a 10.00V shows consumption of 0.000A (i.e. < 0.0005A, very low resistance), and measured voltage is about 9.95V.
Now I also have a homemade hot plate. Hooking this up directly to the power supply, at a constant 10.00V it draws about 2.000A (hence 20W total).
However if I now connect my measurement wires to measure the voltage between the black and red terminals, the device no longer reads 9.95V but rather 9.70V.
This is well outside the stated error margins of the measurement device (it claims ~0.3%, or 30 mV).
I'm thinking the most likely culprit is that at a higher load, the power supply's display is lying. Is that indeed the likeliest possibility? What other sources of error can I look for? I would tend to trust the measuring device as it's much more expensive and also appears to be high quality. But what I trust least of all is my circuit building skills!
Update
If I connect the datalogger directly at the power supply terminals, I get the 9.96V reading.
If I connect the datalogger to the end of the alligator clips coming out of the power supply leads, I get the same ~9.96V reading:
To me this indicates the extremely low resistance of the circuit up to this point.
- Now if I connect my hot plate wires later on the alligator clip, keeping all else the same, the reading on the data logger now drops to 9.688V:
- I get this same ~9.67V reading if I try the following connector options too:
a. Join the data logger and hot plate wire, clip in the middle:
b. Clip the alligator clip to the end of the hot plate wire, and wrap the data logger wire a bit before:
What puzzles me is why the sole act of hooking up the hot plate wires is changing the voltage here. The resistance of the measuring circuit is very small. As I understand it, this forms a parallel circuit -- one going to the data logger, and another going to the hot plate -- where each branch should have the same voltage as the whole circuit, namely 9.96V as measured at the power supply leads.