\$\begingroup\$

In a setup, assuming a function generator has 50 Ohm output impedance and a coaxial cable has 50 Ohm characteristic impedance. Therefore to prevent reflection of a signal a 50 Ohm termination is used. But for many people I know they dont use 50 Ohm termination for basic use(non RF) and they also use 1X probe with scope's 1X setting.

And if we use a 50 Ohm termination then our signal's amplitude halves. How is this problem compensated? By setting the scope to 2X? How about the probe? 1X or 10X?

What is the proper probe/scope and termination resistor setup depending on the signal frequency? Can we roughly make a category? I want to learn the proper measuring technique with scope and probe depending on the waveform and frequency. Since these are based on experience I cant find a compact info about it.

\$\endgroup\$
1
  • \$\begingroup\$ Btw this guy says here:youtu.be/tClE8s6RZdg?t=143 that he observes twice the signal set by the function generator if he doesn't use the 50 Ohm termination. But I see exactly the same signal without a 50 Ohm termination with my function generator ans scope 1X probe 1X setting at 1kHz. \$\endgroup\$
    – user1999
    Commented Sep 7, 2018 at 16:44

3 Answers 3

7
\$\begingroup\$

Ideally, an oscilloscope would give you an option between 1M ohm input resistance (in parallel with perhaps 15pf capacitance) and a 50-ohm input termination - often cautioned not to apply too much power, lest it over-heat. This 50-ohm internal termination is quite vulnerable to burning up into an open-circuit, so it is wise to do a measurement check to see that it is still there, and still 50-ohms.

Many 'scopes give you no option - the default 1M ohm input applies. In this case find a BNC "T", and add a 50-ohm termination right at the 'scope input. Use the 1X scale, and use 50-ohm coax to connect to your signal source, not the 1X probe. Not as good as the internal termination described above, (a short unterminated section often remains), but its about the best you can do. Don't forget that its there - when you go to measure that +24V DC supply, you may smell smoke.
BNC T with 50 ohms Most 1X/10X probes have cable impedance higher than 50 ohms between probe tip and BNC connector, so don't use a probe in a 50-ohm measuring system to make a careful amplitude measurement.

A good function generator takes care to drive its output through a 50-ohm resistance, so its open-circuit output should be twice as high as its 50-ohm terminated output. It is common for function generator outputs to be 20V p-p open-circuit, and 10V p-p when terminated with 50 ohms. Such a generator might say in its manual, "Will deliver 10v p-p to a 50 ohm load".
A source calibrated to deliver 0 dBm power will deliver that power into a proper load (usually 50 ohm) - of course it delivers no power to an open-circuit, and very little more to a 1M 'scope input. Your oscilloscope (set to 1X) might make an RMS amplitude measurement - power is simply \$ \frac {{V_{rms}}^2}{50} \$ when you've got that BNC-T with 50 ohm termination attached.

\$\endgroup\$
4
  • 1
    \$\begingroup\$ You could also use a feed-through 50 Ohm load, as opposed to a T + terminator. Something like this \$\endgroup\$ Commented Sep 7, 2018 at 21:53
  • 1
    \$\begingroup\$ @BlairFonville Good point - a 50-ohm feed-through might go to higher frequencies. Didn't have one to photograph. The one shown is four 220 ohm in parallel with 560. A slug-type 50-ohm resistor inside a feed-through terminator would benefit from having less inductance. \$\endgroup\$
    – glen_geek
    Commented Sep 7, 2018 at 22:24
  • \$\begingroup\$ I didnt get "source calibrated to deliver 0 dBm power" part. Do you mean the function generator can deliver inky 1 mW? \$\endgroup\$
    – user1999
    Commented Sep 8, 2018 at 10:48
  • \$\begingroup\$ @newage2000 If you needed a 0 dBm sine wave from your function generator, you'd connect it with 50-ohm coax to your 'scope, with a 50-ohm terminator attached. Adjust function generator amplitude for 0.633V peak-to-peak on the 'scope (X1). Or 0.224V RMS. Then 1mW is delivered to the terminator. You assume that the 1M-ohm 'scope makes a voltage measurement, and loads the terminator insignificantly. \$\endgroup\$
    – glen_geek
    Commented Sep 8, 2018 at 13:27
4
\$\begingroup\$

I simple terms:

Your default everyday setup with a low-mid range oscilloscope should always be to use 1M input impedance probes and scope set to 10X if available. This setup is non-invasive, and gives you good bandwidth.

If you are a beginner, you can basically stop reading here.

1X mode is meaningful basically only when you need to measure voltages that are too low for 10X mode. Which is never. Almost.

50ohm input mode is only used in certain more advanced cases, where you want to push the bandwidth or have the scope simulate a missing load to avoid reflections. If you connect a signal generator straight to the scope using a coax with no other load on the coax, then 50ohm mode is appropriate. If it is connected usign a probe, then you shall use 1M mode. If you connect the signal generator to a load, and hook up the scope halfway on the coax using a T-connctor, then 1M is the appropriate mode. 50ohm is not "better" than 1M. It is "different", and shall only be used when appropriate.

On a side note, if 10X mode does not offer sufficient bandwidth for a measurement you need to make, and you can't afford high frequency active probes, then there is a neat trick you can do with the 50ohm mode. By using 50ohm mode, a 50ohm coaxial cable, and a series resistor at your measurement point, you can get the bandwidth of an active probe. With the disadvantage of lower impedance. But for digital signals this is often acceptable. The method is explained here: https://www.youtube.com/watch?v=9GqfZMcrAFY

\$\endgroup\$
2
\$\begingroup\$

And if we use a 50 Ohm termination then our signal's amplitude halves. How is this problem compensated?

There is no "problem" at all. Signal is whatever it is, and scope probe has nothing to do with the reality of it. If you don't use 50-Ohm termination, the signal "doubles" as compared to what the signal generator dial assumes (although there could be scaling options depending on assumed termination setup. But is done for dial convenience only). If you load the other cable end with 50-Ohm terminator, the voltage will be half of that. It is what it is, and you should always take measurements with high-impedance probe (>> than 50 Ohms). There are coax probes with 2K input impedance, or even with 500-Ohms, which are still qualified as "non-invasive" for 50-Ohms signaling environment, technically.

There are however situations when you want to use INTERNAL 50-Ohm IMPEDANCE input mode that can be found in modern high-speed digital scopes, for direct connection to coax cables. In this case everything gets matched automatically, yet the signal is still what is is, and no "2x correction" is necessary.

\$\endgroup\$